#1
|
|||
|
|||
BICKEL Gamma Ray Scintillometer
I am in the process of restoring a client's BICKEL "Algor" Gamma Ray Scintillometer and have reversed engineered the schematic. I am presently awaiting new replacement parts. Will also improve on the circuit topology; specifically the digital counter and the metering pulse shaping circuits. The photomultiplier tube and the NaI(t) crystal in the probe unit are in pristine condition and very sensitive. I would like to contact anyone that has one of the BICKEL units to discover any extant documentation on its operation and a more detailed history.
Though it has some clever design features, it is however a basic gamma ray scintillometer with very limited spectral discrimination capabilities. Its main detection facilities include a 3 in. NaI(t) crystal coupled to RCA6342A Photomultiplier tube. However the probe unit has NO anti-magnetic shield typically made of monel or mu-metal that must surround the photomultiplier tube. Without such a shield the large photomultiplier tabe is susceptible to weak magnetic fields in its vicinity including the earth's magnetic field. It's omission would render any readings from the unit unreliable. The crystal is a standard unit manufactured by Teledyne Brown Engineering. Oddly, despite some of the claims made by others the unit is a crudely made home brew unit. It has a somewhat more complex circuit topology than units made by others such as PRI, Ludlum, Bicron, McFar, Eberline, Detectron, Sopris, etc. It is similar to units manufactured since the late 1940- to 1950's. In terms of sensitivity, it is equivalent to the legendary PRI-111C or a PRI-118 which are available unrestored from time to time on EBAY.. We have been restoring, upgrading, and refurbing vintage Radiation detection equipment for over 30 years. |
#2
|
||||
|
||||
Try to send an email or PM to Ducati250, but i am not sure if he still reads the forum
http://www.longrangelocators.com/for...ad.php?t=19048. Also can you attach here some photos from the device??, i like to collect them Regards
__________________
Geo |
#3
|
|||
|
|||
Bickel photos
Geo,
Thanks for the heads up on Ducati. I believe he sold that unit to my client and it appears identical vis-Ã*-vis the photos he posted. I am unable to upload photos of the unit at this time as they are not on a url but on our system. I see no provision on this site to upload pictures from our system onto a posting Once the unit is fully repaired and upgraded I may put photos and details on a web page and will post a link to it. Despite previous assertions by J.Player, the PMT and Scintillator crystal are fairly standard fare and are manufactured domestically in the USA. The leading Manufacturer of PMT's today is Hamamatsu in Japan. There also several manufacturers of Scintillator crystals such as Saint Gobain, et. al. PMT's and scintillators are used extensively in nuclear research and there are literally thousands used to detect emissions of sub-atomic particles in the LHC (Large Hadron Collider) in Geneva Switzerland. The Bickel units are no more sensitive than other scintillator units that have the same spec photomultiplier tube and same sized NaI(t) crystal. In fact such units have been available even to the general public since the 1950's and have been used to find radioactive ore fields, oil, precious metal ores, and even dinosaur bones. |
#4
|
|||
|
|||
Hi zidex, I was wondering, has some kind of discrimination or it's just a (sensitive) counter?
|
#5
|
||||
|
||||
@zidex666: "I see no provision on this site to upload pictures from our system onto a posting "
To upload pictures you should go in Reply down to Options: Additional Options/Attach Files
__________________
Global capital is ruining your life? You have right to self-defence! |
#6
|
|||
|
|||
Counts & Discrimination
Scintillometers are not counters as such. Geiger counters are, as they count the number of beta particles and in some instances low energy gamma rays that are detected by a Geiger-muller tube, They were in the past measuring CPM or counts per minute. However nowadays their digital meters or analog indicators are calibrated into Sieverts or other units.
Scintillators were traditionally calibrated in MR/HR or milli-roentgens per hour. They do not count ionized radioactive particles per se as with a Geiger Counter but rather measure gamma-ray intensity using various standard energy measurement units. It is also generally true that the bigger the Geiger-Muller tube is the more sensitive it becomes. The same principle holds true for a scintillator. The bigger the Scintillation crystal the more sensitive the unit becomes. It's much like catching butterflies in a open field with either a net with a 12 inch opening or with a net with a 36 inch opening. Which net do you think will catch the most butterflies if given an equal amount of time? However one reaches a point of diminishing returns when increasing the size or cross section of a scintillation crystal. Namely "background" radiation. As you increase the sensitivity of a scintillator you increase the level of background radiation proportionately and any signal or signature of interest gets buried in the noise. Kinda trying to listen to an AM broadcast station during the day as you travel away from it. Its signal eventually gets buried in the noise. Naturally occuring ubiquitous thorium radiation in soils particularly in the USA Midwest, radon, cosmic rays, solar flares, hospitals, technicium injected patients, and ionospheric disturbances all contribute to the natural background radiation and limits scintillation accuracy. Remember there are 3 basic types of ionizing radiation; Alpha, Beta, and Gamma. Alpha radiation consists of sub-atomic particles and are really helium nuclei. Alpha particles can be blocked or stopped with several sheets of paper and travel about 10 inches or so. Beta radiation consists of either electrons or negative matter positrons. they are particles and can be stopped by several layers of lead shielding. They can travel several hundred feet until they are absorbed into other matter. Gamma radiation on the other hand is not particle radiation but a ray but on a higher energy level than Light, Ultraviolet, and X-Rays. Gamma rays penetrate matter and can penetrate a lead shield millions of miles thick. Gamma ray bursts have been detected from black hole events in distant galaxies and within the Milky way. The only discrimination that's possible from a scintillator other than measuring the intensity of the gamma radiation is to measure its energy bandwidth which can range from several thousand electron volts to several million (MEV) with sophisticated processing electronics such as multi-channel analyzers or by carefully selected analog circuits. However, analog circuits are not very reliable nor very accurate and only yield very coarse indications of detected gamma ray energy levels. The ability to discern differences between gamma-ray energy levels (but not necessarily intensity) is "discrimination". It is true that certain non-stable elements emit gamma radiation with various intensities but also may emit, for instance for an element X, gamma rays at energy levels of 56KEV, 137KEY, and 2.2MEV simultaneously. Those levels and intensity markers can be used to identify element X and that phenomena is what constitutes some of its "signature." The analysis of those energy levels, intensity, and signature is what is called gamma-ray spectroscopy. However what else is needed for a comprehensive element signature analysis is the inclusion of other signature emissions given off by an element. This includes the intensity and amount of alpha and beta particle radiation which cannot be detected by scintillators. It appears that the Bickel unit can be roughly calibrated to "discriminate" a certain small bandwidth of detected gamma-ray energies while rejecting others via simple pulse shaping and basic electronic attenuation networks. Whether that is sufficient to accurately identify any isotope or any element below Pb (lead) besides radioactive K (potassium) is questionable or at least suspect at best. Most elements below (lead) have isotopes but are not necessarily radioactive and rarely emit gamma, alpha or beta radiation. If they do, the intensity is so small or the half-life is so short that field equipment would not have the sensitivity to detect it. Furthermore, the percentage of unstable isotopes versus a stable isotope in elements below lead in the periodic table is so small that trying to identify it with a scintillator would be difficult if not impossible. Case in point is measuring the proportion of Carbon 14 in a sample of interest for carbon dating. It would require very sensitive and sophisticated laboratory equipment and chemical reagents to measure. Ditto with trying to identify heavy water from normal water using scintillation. Hope this helps somewhat. Regards |
#7
|
|||
|
|||
Bickel Photos
Here's some photos of the Bickel Unit.
Zidex666 Sorry seems this site wants me to reduce the size of the jpg images. Will re-post later. Last edited by zidex666; 02-23-2016 at 03:44 AM. Reason: spelling correction |
#8
|
||||
|
||||
Zidex ...thanks for the seminar
__________________
Geo |
#9
|
|||||
|
|||||
Quote:
Anyway, my true question was : discrimination or intensity? Quote:
Quote:
Gammas, have a relatively weak interaction with matter in comparison with the alpha+beta, but interacts, and by this interaction we have absorption, attenuation, detection,etc About these exo-terrestrial gamma sources, the gamma photons travel through (almost) empty space, not through dense matter. Quote:
Quote:
|
#10
|
|||
|
|||
Bickel Detector Photos
Please find attached Photos of the Bickel unit including the Probes internals.
|
#11
|
||||
|
||||
Thank you very much
BTW... why there is isolation tape around the tubes???
__________________
Geo |
#12
|
|||
|
|||
Bickel Photos & PMT/Crystal taping
The black tape serves three main functions:
It provides shielding against light entering the photomultiplier tube. (You only want light or photons emitted by the crystal to enter the photo the top of the photomultiplier tube.) It crudely binds the NaI(t) crystal holder to the top of the photomultiplier tube. Unfortunately the Bickel probe did not have any optical grease between the photomultiplier and the crystal holder when it was disassembled. This is bad practice as stray humidity will fog up the junction between the photomultiplier and the crystal. It testifies that the unit's probe was constructed in a home-brew Rub-Goldberg fashion. Thirdly, on top of the first layer of tape, a crude and cheap lead shield made of a cut-up sheet of photographic "FilmShield" was wrapped around the Photomultiplier tube. Whether this simple lead shielding had any significant affect while prospecting or searching for certain radioactive isotopes is uncertain. If anything it may provide limited shielding of neutron radiation. Perhaps the most significant shortcoming in the probe's construction is the complete absence of any anti-magnetic shielding. Photomultiplier tubes are extremely susceptible to stray magnetic fields and can seriously affect their operation and accuracy. In fact even the earth's magnetic field can be a problem in rendering accurate readings. Inspection of the inside of the probe's outer casing revealed no such shielding. Anti-magnetic shields are encased around photomultiplier tubes and are typically made of Monel or mu-metal and are fairly ubiquitous in scintillator probes. In general, the lack of anti-magnetic shield in the probe makes one question the Bickel unit's efficacy in actually identifying radioactive isotopes. Notice that in one of the photos that the tape was removed from the PMT and the crystal holder. The PMT is of USA (RCA) manufacture as is the NaI(t) hermetically sealed crystal holder made by Teledyne. NaI(t) crystals are extremely hygroscopic particularly when exposed to humid air. If exposed in an environment with any humidity they quickly turn yellow and opaque rendering them useless. NaI(t) is also extremely toxic. Crystal holders produced in the 1950's often turned the crystals yellow and opaque after two to three years because they were sealed using rubber O-rings which quickly deteriorated and allowed water vapor inside the crystal housing. Later in 1960's onward they were hermetically sealed by almost all manufacturers. Notice that in the photo where the crystal holder is separated from the PMT, the crystal if absolutely clear and Not opaque. We will rebuild the entire probe unit and bring it up to modern specs. Hope this helps. Regards. Last edited by zidex666; 02-25-2016 at 04:25 PM. Reason: spelling error |
#13
|
|||
|
|||
thanks for the update zidex.
Quote:
|
#14
|
||||
|
||||
In response to Bill512:
Yes there are some scintillator probes that can be connected to certain vintage counting units made by such firms as Ludlum, Bicron, Eberline, Eltronics, etc.. and those unit are calibrated in CPM. However most vintage standalone Scintillators are calibrated in the now obsolete MR/HR. There is no standard way to convert CPM to MR/HR conveniently or accurately. It's essentially comparing apples to oranges. Hence new measuring standards are gradually being phased in. The point was that normally the output of a scintillator probe basically measures gamma flux intensity and does not rely on simple counting circuits as in Geiger Counters. so the original posting stands. Quote:
This is only true in an conceptual idealized isotropic environment. Without going into the subtleties of Shannon's communications theory and simply utilizing practicle real world effects, signal to noise ratios are NOT fixed but independently variable. Either the source may be amplitude varying or the Background noise maybe be amplitude and time varying. In either or both cases the signal to noise ration is affected. If the background noise is of greater amplitude than the source signal, the source is swamped out by the background noise. The result is that the signal to noise ratio changes. The examples noted in the original posting are not merely anecdotal but factual in real world environments. Quote:
Quote:
But, thanks for the kudos. Quote:
Regards. Last edited by zidex666; 02-25-2016 at 06:34 PM. Reason: Spelling errors |
#15
|
|||
|
|||
Quote:
It does NOT wrap around the Crystal housing! Regards! Last edited by zidex666; 02-25-2016 at 07:37 PM. Reason: spelling |
#16
|
|||
|
|||
More Bickel unit photos
More photos:
|
#17
|
|||
|
|||
General observations of the Bickel Unit panel controls
Looking at the front panel controls the following is observed:
The two smaller meters are over kill. One on the left center indicates the unit's current draw and the small one on the right merely indicates volts or battery condition. Those measurements could easily be combined in one meter. Great visual effect though with an unnecessary surplus of meters. Knobs, switches etc. are standard Radio Shack items. Cheesy off center mounted housing of the fuse holder. The "subnormal" meter is likely utilized to indicate background noise and may be utilized to indicate what affect that the pulse shaping circuits can be utilized to attenuate it. The "abnormal" meter is likely to read the amplitude and intensity of the gamma radiation. In fact both of these functions (subnormal and abnormal) can be combined in one large meter. It does however add to its visual appeal for some. The gothic lettering may be cheesy to some but as Dr. Bickel was a Nazi Rocket Scientist the choice of the gothic press-on lettering on the meter and panel faces makes sense. Depending on the client's choice we may either change the lettering or leave it as is to preserve the unit's originality. |
#18
|
|||
|
|||
Quote:
However is true that the output of a scintillation tube is pulsed voltage directly related to the initial product of the NaI(T) photons? (photoelectric superimposed in compton). Quote:
Now, lets see the more realistic cases. Any time variance of the sources flux, has effect on the noise to signal ratio but this has nothing to do with any detector. Let the ratio be variable all the time, with whatever function you like. The point is that the "ultimate" (barrier) signal to noise ratio is a property of the space it's self, and not instrumentation related. As you make an instrument more and more sensitive there is a point beyond that any improvement in terms of sensitivity is useless. This is not because the signal then is buried in noise as in your statement: "..As you increase the sensitivity of a scintillator you increase the level of background radiation proportionately and any signal or signature of interest gets buried in the noise.." This is simply because, beyond this point you amplify the signal as well the noise - so, no real benefit except signal scaling. sorry, but these "millions of miles of lead shield" was to shield the gamma radiation of a cosmic event in the order of magnitude of the bing bang? anyway,many thanks for the Bickel info and the nice talking. this instrument was mysterious, not to say (almost) mythical... Bill |
#19
|
||||
|
||||
After your discussion I feel to be more irrelevant but thanks again
__________________
Geo |
#20
|
|||
|
|||
|
#21
|
|||
|
|||
Vintage units similar to the Bickel Detector
For comparison here are some Photos of a Vintage scintillator (gamma ray detector) unit similar in function and sensitivity the BICKEL unit. The PRI-118 was manufactured and sold during the mid to late 1950's. It utilized sub-miniature vacuum tubes developed during WW2. The appear from time to time on EBAY.
Regards Last edited by zidex666; 02-26-2016 at 10:25 PM. Reason: Spelling & additional photos |
#22
|
|||
|
|||
Photo of a bad NaI(t) scintillator crystal
Notice the canister below with the yellow crystal. The rubber O ring seal has leaked allowing moisture into the crystal housing. Crystal units constructed in the 1950 had this problem. They typically lasted only 2 to 3 years. Once yellowed they become useless unless the crystal is removed and replaced in a moisture free environment. Sometimes they can be re-polished. In the 1960' on, manfacturers started to hermetically seal them so that they would last longer.
As mentioned in previous posts by me, NaI(t) is extremely toxic and hygroscopic. It corrodes quickly when exposed to any moisture in the air . Ggreat care and precautions most be made when replacing or re-polishing it once it has yellowed. Once yellowed and no longer transparent, the crystals are useless. The crystal along with the photomultiplier are the heart of any scintillator unit. The rest are simply electronics that process the output from the photomultiplier tube. In addition to Nai(t) other materials such as special plastics and other compounds are used for monitoring certain bandwidths of the X-ray/Gamma spectrum are used today. The myth that only the most sensitive scintillation crystal and/or photomultiplier tubes are only available from Germany is inaccurate and false. Both are now manufactured worldwide by many firms including some in Russia and China. Regards Last edited by zidex666; 02-26-2016 at 10:50 PM. Reason: spelling |
#23
|
||||
|
||||
Hello.
Today on ebay we can find scintillation crystals and photomultiplier tubes easy and at low prices. How difficult or easy is to construct a new device like Dr Brickel's ? For example Dr brickel was using a lot of components to make a digital counter and now it is fixed (i mean all components and display at same small case) at very low price. Except the protectiob for good shielding are there other things who makes the construction unposible or unworkable? Regards
__________________
Geo |
#24
|
|||
|
|||
Quote:
Yes there are on Ebay. Some really good ones are available from some Russian sellers. I suggest that you get a 6619 photomultiplier tube to mate with a 1 inch diameter NaI(t) crystal as a first project. Another idea is to get a vintage PRI-111 or PRI-111b scintillator or an alternative unit off EBAY and restore it. Making Bickel's digital counter is unnecessary simply because you can buy panel mounted digital counter modules on EBAY for less than $20 USD. It works kind of like a simple kitchen digital count-down timer. All that counter does is to count "seconds" backwards to shut off an external electromechanical paper chart unit than can be connected to the Bickel unit. There is simply NO way that the Bickel Counter, no matter what its count display, can accurately identify an individual isotope unless a scalar reading with an external chart recorder is utilized. Even then it's only a guess at best. It merely counts time in Seconds not isotope number. Furthermore, it is extremely improbable that the Bickel unit can identifiy any gold isotope by reading some mysterious number on the counter display. Gold in nature is stable and any gold isotope that is radioactive can only be so if it is "synthetic." That means that it must be strongly irradiated with radioactivity and in essentially laboratory conditions. Additionally any such synthetic radioactive gold isotopes have comparatively short half-lives. The only things that the Bickel unit along with other NaI(t) based scintillometers detect and are most useful for are radioactive sources emitting gamma/xray radiation in the field. Primarily decay products from Thorium and Uranium as they eventually decay into lead. Do a search on Uranium and Thorium decay products (example radium. etc.) to learn how Uranium or Thorium eventually transmute into Lead. Also, look up the half-life of the individual elements as they decay into lighter elements in the decay chain. Finally, there is NO magic or obscure Nazi Science manifest in the Bickel unit. It's a rather conventional Scintillator design with an unnecessary bunch of extra bells and whistles, fancy knobs and unnecessary meters. In order to identify ore fields that were found using the Bickel unit, other extant scintillator at the time could have given equivalent if not superior performance. Without a good knowledge of the geology of the terrain, the out-croppings, and other envronmental factors, which the Bickel may certainly have had, a scintillator will not directly identify oil, gold, dinosaur bones, precious metals, or treasures. The exception maybe if someone buried a nuclear warhead or radioactive uranium mine tailings. Any NaI(t) based scintillator on the other hand CAN detect long range radioactive (trans-uranium and trans-thorium) ore bodies beneath the surface. Identifying the ore body and its constituents, one can properly identify any other elements such as gold, oil, or other things requires an intimate knowledge of geology, chemical testing, borehole sampling, and other factors in addition to a scintillator. See below some pictures of common vintage scintilltors we've restored. FYI: The are some excellent YAHOO interest Groups that are devoted to Geiger Counters and Scintillators. Regards Last edited by zidex666; 02-27-2016 at 11:33 PM. Reason: spelling and syntax |
#25
|
||||
|
||||
Zidex thank you.
Do you think that one device like these can catch buried gold 100 years at depth 4m and weight 500kgr?? Regards
__________________
Geo |
Tags |
bickel, gamma ray, radiation detection, scintillator |
|
|