Note that D1 limits the voltage across the "meter + R24" combination to 0.7V. Without R4, the current through the meter would be: 0.7V / 675 = 1.04mA. This is clearly too high for the 250uA meter. Hence R24 (3k) is used to limit the maximum current to: 0.7 / (3k + 675) = 190uA. The slightly lower then 250uA value stops the meter needle banging against the meter stop when a large signal is received.

All you need to do is replace R24 with a suitable value for your meter. Let's say (for example) your meter is 100uA full-scale deflection (FSD) with 1k5 resistance. A total resistance of 0.7V / 100uA = 7k. Hence, R24 needs to be 7k - 1k5 = 5k5 (or slightly higher). A simple solution would be to replace R24 with a 10k multi-turn preset, and manually adjust it to suit your particular meter.

It should also be obvious that your chosen meter needs to have an FSD of less than 1.04mA.

Thank you very much Qiaozhi. You are a great help.