PDA

View Full Version : BICKEL Gamma Ray Scintillometer


zidex666
02-22-2016, 08:38 PM
I am in the process of restoring a client's BICKEL "Algor" Gamma Ray Scintillometer and have reversed engineered the schematic. I am presently awaiting new replacement parts. Will also improve on the circuit topology; specifically the digital counter and the metering pulse shaping circuits. The photomultiplier tube and the NaI(t) crystal in the probe unit are in pristine condition and very sensitive. I would like to contact anyone that has one of the BICKEL units to discover any extant documentation on its operation and a more detailed history.

Though it has some clever design features, it is however a basic gamma ray scintillometer with very limited spectral discrimination capabilities. Its main detection facilities include a 3 in. NaI(t) crystal coupled to RCA6342A Photomultiplier tube. However the probe unit has NO anti-magnetic shield typically made of monel or mu-metal that must surround the photomultiplier tube. Without such a shield the large photomultiplier tabe is susceptible to weak magnetic fields in its vicinity including the earth's magnetic field. It's omission would render any readings from the unit unreliable. The crystal is a standard unit manufactured by Teledyne Brown Engineering. Oddly, despite some of the claims made by others the unit is a crudely made home brew unit. It has a somewhat more complex circuit topology than units made by others such as PRI, Ludlum, Bicron, McFar, Eberline, Detectron, Sopris, etc. It is similar to units manufactured since the late 1940- to 1950's. In terms of sensitivity, it is equivalent to the legendary PRI-111C or a PRI-118 which are available unrestored from time to time on EBAY..

We have been restoring, upgrading, and refurbing vintage Radiation detection equipment for over 30 years.

Geo
02-22-2016, 08:53 PM
Try to send an email or PM to Ducati250, but i am not sure if he still reads the forum

http://www.longrangelocators.com/forums/showthread.php?t=19048.
Also can you attach here some photos from the device??, i like to collect them

Regards

zidex666
02-22-2016, 10:57 PM
Geo,

Thanks for the heads up on Ducati. I believe he sold that unit to my client and it appears identical vis-Ã*-vis the photos he posted. I am unable to upload photos of the unit at this time as they are not on a url but on our system. I see no provision on this site to upload pictures from our system onto a posting

Once the unit is fully repaired and upgraded I may put photos and details on a web page and will post a link to it.

Despite previous assertions by J.Player, the PMT and Scintillator crystal are fairly standard fare and are manufactured domestically in the USA. The leading Manufacturer of PMT's today is Hamamatsu in Japan. There also several manufacturers of Scintillator crystals such as Saint Gobain, et. al.

PMT's and scintillators are used extensively in nuclear research and there are literally thousands used to detect emissions of sub-atomic particles in the LHC (Large Hadron Collider) in Geneva Switzerland. The Bickel units are no more sensitive than other scintillator units that have the same spec photomultiplier tube and same sized NaI(t) crystal. In fact such units have been available even to the general public since the 1950's and have been used to find radioactive ore fields, oil, precious metal ores, and even dinosaur bones.

Bill512
02-22-2016, 11:18 PM
... it is however a basic gamma ray scintillometer with very limited spectral discrimination capabilities...
Hi zidex, I was wondering, has some kind of discrimination or it's just a (sensitive) counter?

WM6
02-23-2016, 12:56 AM
@zidex666: "I see no provision on this site to upload pictures from our system onto a posting "

To upload pictures you should go in Reply down to Options:

Additional Options/Attach Files

zidex666
02-23-2016, 03:21 AM
Scintillometers are not counters as such. Geiger counters are, as they count the number of beta particles and in some instances low energy gamma rays that are detected by a Geiger-muller tube, They were in the past measuring CPM or counts per minute. However nowadays their digital meters or analog indicators are calibrated into Sieverts or other units.

Scintillators were traditionally calibrated in MR/HR or milli-roentgens per hour. They do not count ionized radioactive particles per se as with a Geiger Counter but rather measure gamma-ray intensity using various standard energy measurement units.

It is also generally true that the bigger the Geiger-Muller tube is the more sensitive it becomes. The same principle holds true for a scintillator. The bigger the Scintillation crystal the more sensitive the unit becomes. It's much like catching butterflies in a open field with either a net with a 12 inch opening or with a net with a 36 inch opening. Which net do you think will catch the most butterflies if given an equal amount of time?

However one reaches a point of diminishing returns when increasing the size or cross section of a scintillation crystal. Namely "background" radiation. As you increase the sensitivity of a scintillator you increase the level of background radiation proportionately and any signal or signature of interest gets buried in the noise. Kinda trying to listen to an AM broadcast station during the day as you travel away from it. Its signal eventually gets buried in the noise. Naturally occuring ubiquitous thorium radiation in soils particularly in the USA Midwest, radon, cosmic rays, solar flares, hospitals, technicium injected patients, and ionospheric disturbances all contribute to the natural background radiation and limits scintillation accuracy.

Remember there are 3 basic types of ionizing radiation; Alpha, Beta, and Gamma.
Alpha radiation consists of sub-atomic particles and are really helium nuclei. Alpha particles can be blocked or stopped with several sheets of paper and travel about 10 inches or so.
Beta radiation consists of either electrons or negative matter positrons. they are particles and can be stopped by several layers of lead shielding. They can travel several hundred feet until they are absorbed into other matter. Gamma radiation on the other hand is not particle radiation but a ray but on a higher energy level than Light, Ultraviolet, and X-Rays.
Gamma rays penetrate matter and can penetrate a lead shield millions of miles thick. Gamma ray bursts have been detected from black hole events in distant galaxies and within the Milky way.


The only discrimination that's possible from a scintillator other than measuring the intensity of the gamma radiation is to measure its energy bandwidth which can range from several thousand electron volts to several million (MEV) with sophisticated processing electronics such as multi-channel analyzers or by carefully selected analog circuits. However, analog circuits are not very reliable nor very accurate and only yield very coarse indications of detected gamma ray energy levels. The ability to discern differences between gamma-ray energy levels (but not necessarily intensity) is "discrimination".

It is true that certain non-stable elements emit gamma radiation with various intensities but also may emit, for instance for an element X, gamma rays at energy levels of 56KEV, 137KEY, and 2.2MEV simultaneously. Those levels and intensity markers can be used to identify element X and that phenomena is what constitutes some of its "signature." The analysis of those energy levels, intensity, and signature is what is called gamma-ray spectroscopy. However what else is needed for a comprehensive element signature analysis is the inclusion of other signature emissions given off by an element. This includes the intensity and amount of alpha and beta particle radiation which cannot be detected by scintillators.

It appears that the Bickel unit can be roughly calibrated to "discriminate" a certain small bandwidth of detected gamma-ray energies while rejecting others via simple pulse shaping and basic electronic attenuation networks. Whether that is sufficient to accurately identify any isotope or any element below Pb (lead) besides radioactive K (potassium) is questionable or at least suspect at best. Most elements below (lead) have isotopes but are not necessarily radioactive and rarely emit gamma, alpha or beta radiation. If they do, the intensity is so small or the half-life is so short that field equipment would not have the sensitivity to detect it. Furthermore, the percentage of unstable isotopes versus a stable isotope in elements below lead in the periodic table is so small that trying to identify it with a scintillator would be difficult if not impossible. Case in point is measuring the proportion of Carbon 14 in a sample of interest for carbon dating. It would require very sensitive and sophisticated laboratory equipment and chemical reagents to measure. Ditto with trying to identify heavy water from normal water using scintillation.

Hope this helps somewhat.

Regards

zidex666
02-23-2016, 03:40 AM
Here's some photos of the Bickel Unit.

Zidex666

Sorry seems this site wants me to reduce the size of the jpg images.
Will re-post later.

Geo
02-23-2016, 06:30 AM
Zidex ...thanks for the seminar :thumb:

Bill512
02-23-2016, 11:24 PM
Scintillometers are not counters as such.
Scintillators were traditionally calibrated in MR/HR or milli-roentgens per hour. They do not count ionized radioactive particles per se as with a Geiger Counter but rather measure gamma-ray intensity using various standard energy measurement units.

true, partially,there are some scintillation probes for Geiger's (with CPM "units").
Anyway, my true question was : discrimination or intensity?


However one reaches a point of diminishing returns when increasing the size or cross section of a scintillation crystal. Namely "background" radiation. As you increase the sensitivity of a scintillator you increase the level of background radiation proportionately and any signal or signature of interest gets buried in the noise.

At a given point of space and for a specific "bandwidth", the signal to noise ratio (a radiation source vs background radiation) is fixed , regardless if you carry a scintillation meter ,a Geiger or nothing.

Alpha particles can be blocked or stopped with several sheets of paper and travel about 10 inches or so.
Beta radiation consists of either electrons or negative matter positrons. they are particles and can be stopped by several layers of lead shielding. They can travel several hundred feet until they are absorbed into other matter. Gamma radiation on the other hand is not particle radiation but a ray but on a higher energy level than Light, Ultraviolet, and X-Rays.
Gamma rays penetrate matter and can penetrate a lead shield millions of miles thick. Gamma ray bursts have been detected from black hole events in distant galaxies and within the Milky way.
False 100% ,especially the beta and gamma statements. millions of miles of lead shield ?
Gammas, have a relatively weak interaction with matter in comparison with the alpha+beta, but interacts, and by this interaction we have absorption, attenuation, detection,etc
About these exo-terrestrial gamma sources, the gamma photons travel through (almost) empty space, not through dense matter.

The only discrimination that's possible from a scintillator other than measuring the intensity of the gamma radiation is to measure its energy bandwidth which can range from several thousand electron volts to several million (MEV) with sophisticated processing electronics such as multi-channel analyzers or by carefully selected analog circuits. However, analog circuits are not very reliable nor very accurate and only yield very coarse indications of detected gamma ray energy levels. The ability to discern differences between gamma-ray energy levels (but not necessarily intensity) is "discrimination".

It is true that certain non-stable elements emit gamma radiation with various intensities but also may emit, for instance for an element X, gamma rays at energy levels of 56KEV, 137KEY, and 2.2MEV simultaneously. Those levels and intensity markers can be used to identify element X and that phenomena is what constitutes some of its "signature." The analysis of those energy levels, intensity, and signature is what is called gamma-ray spectroscopy. However what else is needed for a comprehensive element signature analysis is the inclusion of other signature emissions given off by an element. This includes the intensity and amount of alpha and beta particle radiation which cannot be detected by scintillators.
thanks for the big post, I know the limitations of gamma ray spectroscopy- it's not magic but applied science.

It appears that the Bickel unit can be roughly calibrated to "discriminate" a certain small bandwidth of detected gamma-ray energies while rejecting others via simple pulse shaping and basic electronic attenuation networks. Whether that is sufficient to accurately identify any isotope or any element below Pb (lead) besides radioactive K (potassium) is questionable or at least suspect at best...
thanks...

zidex666
02-25-2016, 09:48 AM
Please find attached Photos of the Bickel unit including the Probes internals.

Geo
02-25-2016, 11:27 AM
Thank you very much:)

BTW... why there is isolation tape around the tubes???

zidex666
02-25-2016, 04:04 PM
The black tape serves three main functions:

It provides shielding against light entering the photomultiplier tube. (You only want light or photons emitted by the crystal to enter the photo the top of the photomultiplier tube.)

It crudely binds the NaI(t) crystal holder to the top of the photomultiplier tube.
Unfortunately the Bickel probe did not have any optical grease between the photomultiplier and the crystal holder when it was disassembled. This is bad practice as stray humidity will fog up the junction between the photomultiplier and the crystal. It testifies that the unit's probe was constructed in a home-brew Rub-Goldberg fashion.

Thirdly, on top of the first layer of tape, a crude and cheap lead shield made of a cut-up sheet of photographic "FilmShield" was wrapped around the Photomultiplier tube. Whether this simple lead shielding had any significant affect while prospecting or searching for certain radioactive isotopes is uncertain. If anything it may provide limited shielding of neutron radiation.

Perhaps the most significant shortcoming in the probe's construction is the complete absence of any anti-magnetic shielding. Photomultiplier tubes are extremely susceptible to stray magnetic fields and can seriously affect their operation and accuracy. In fact even the earth's magnetic field can be a problem in rendering accurate readings. Inspection of the inside of the probe's outer casing revealed no such shielding. Anti-magnetic shields are encased around photomultiplier tubes and are typically made of Monel or mu-metal and are fairly ubiquitous in scintillator probes. In general, the lack of anti-magnetic shield in the probe makes one question the Bickel unit's efficacy in actually identifying radioactive isotopes.

Notice that in one of the photos that the tape was removed from the PMT and the crystal holder. The PMT is of USA (RCA) manufacture as is the NaI(t) hermetically sealed crystal holder made by Teledyne.

NaI(t) crystals are extremely hygroscopic particularly when exposed to humid air. If exposed in an environment with any humidity they quickly turn yellow and opaque rendering them useless. NaI(t) is also extremely toxic. Crystal holders produced in the 1950's often turned the crystals yellow and opaque after two to three years because they were sealed using rubber O-rings which quickly deteriorated and allowed water vapor inside the crystal housing. Later in 1960's onward they were hermetically sealed by almost all manufacturers. Notice that in the photo where the crystal holder is separated from the PMT, the crystal if absolutely clear and Not opaque.

We will rebuild the entire probe unit and bring it up to modern specs.

Hope this helps.

Regards.

Bill512
02-25-2016, 05:42 PM
thanks for the update zidex.


...Thirdly, on top of the first layer of tape, a crude and cheap lead shield made of a cut-up sheet of photographic "FilmShield" was wrapped around the Photomultiplier tube. Whether this simple lead shielding had any significant affect while prospecting or searching for certain radioactive isotopes is uncertain. If anything it may provide limited shielding of neutron radiation...

about this crude lead thin shield, it is wrapped also all around the NaI(T) crystal??

zidex666
02-25-2016, 06:30 PM
In response to Bill512:

Yes there are some scintillator probes that can be connected to certain vintage counting units made by such firms as Ludlum, Bicron, Eberline, Eltronics, etc.. and those unit are calibrated in CPM. However most vintage standalone Scintillators are calibrated in the now obsolete MR/HR. There is no standard way to convert CPM to MR/HR conveniently or accurately. It's essentially comparing apples to oranges. Hence new measuring standards are gradually being phased in. The point was that normally the output of a scintillator probe basically measures gamma flux intensity and does not rely on simple counting circuits as in Geiger Counters. so the original posting stands.

" At a given point of space and for a specific "bandwidth", the signal to noise ratio (a radiation source vs background radiation) is fixed , regardless if you carry a scintillation meter ,a Geiger or nothing"



This is only true in an conceptual idealized isotropic environment. Without going into the subtleties of Shannon's communications theory and simply utilizing practicle real world effects, signal to noise ratios are NOT fixed but independently variable. Either the source may be amplitude varying or the Background noise maybe be amplitude and time varying. In either or both cases the signal to noise ration is affected. If the background noise is of greater amplitude than the source signal, the source is swamped out by the background noise. The result is that the signal to noise ratio changes. The examples noted in the original posting are not merely anecdotal but factual in real world environments.

False 100% ,especially the beta and gamma statements. millions of miles of lead shield ? Gammas, have a relatively weak interaction with matter in comparison with the alpha+beta, but interacts, and by this interaction we have absorption, attenuation, detection,etc About these exo-terrestrial gamma sources, the gamma photons travel through (almost) empty space, not through dense matter.


Not false! However your statement: "Gammas, have a relatively weak interaction with matter in comparison with the alpha+beta, but interacts, and by this interaction we have absorption, attenuation, detection, etc."



This is generally correct. However any interaction depends on the gamma flux intensity. The greater the intensity or energy flux the greater the interaction with matter per unit volume. Case in point is that ionizing radiation, gamma, x-ray radiation in a nuclear explosion would make even the efficacy of massive lead shielding in a blast zone moot even if thermionic forces don't incinerate any of the lead shielding. The point here in my original posting was to basically differentiate between the various kinds of ionizing radiation for the laymen and the instrumentalities used to detect them. Therefore what's the point of your comment? My posting was not meant to be a comprehensive treatise on nuclear or quantum physics. What I had in my posting stands!

But, thanks for the kudos.

I know the limitations of gamma ray spectroscopy- it's not magic but applied science.


Absolutely agree! It's indeed not magical but applied science.

Regards.

zidex666
02-25-2016, 06:48 PM
about this crude lead thin shield, it is wrapped also all around the NaI(T) crystal?? No. It's width is not wide enough to envelope the entire glass housing of the PMT. It does however, short that it is widthwise, wrap around only the PMT. It's presence though is puzzling. The only reason that would make sense is that it is to shield it from a large flux of neutron radiation. Even so in real world environments lead shielding on a PMT is unnecessary. More important for PMT's is anti-magnetic radiation shielding.

It does NOT wrap around the Crystal housing!

Regards!

zidex666
02-25-2016, 07:05 PM
More photos:

zidex666
02-25-2016, 07:31 PM
Looking at the front panel controls the following is observed:

The two smaller meters are over kill. One on the left center indicates the unit's current draw and the small one on the right merely indicates volts or battery condition. Those measurements could easily be combined in one meter. Great visual effect though with an unnecessary surplus of meters.

Knobs, switches etc. are standard Radio Shack items.

Cheesy off center mounted housing of the fuse holder.

The "subnormal" meter is likely utilized to indicate background noise and may be utilized to indicate what affect that the pulse shaping circuits can be utilized to attenuate it.

The "abnormal" meter is likely to read the amplitude and intensity of the gamma radiation.

In fact both of these functions (subnormal and abnormal) can be combined in one large meter. It does however add to its visual appeal for some.

The gothic lettering may be cheesy to some but as Dr. Bickel was a Nazi Rocket Scientist the choice of the gothic press-on lettering on the meter and panel faces makes sense. Depending on the client's choice we may either change the lettering or leave it as is to preserve the unit's originality.

Bill512
02-25-2016, 08:53 PM
In response to Bill512:

Yes there are some scintillator probes that can be connected to certain vintage counting units made by such firms as Ludlum, Bicron, Eberline, Eltronics, etc.. and those unit are calibrated in CPM. However most vintage standalone Scintillators are calibrated in the now obsolete MR/HR. There is no standard way to convert CPM to MR/HR conveniently or accurately. It's essentially comparing apples to oranges. Hence new measuring standards are gradually being phased in. The point was that normally the output of a scintillator probe basically measures gamma flux intensity and does not rely on simple counting circuits as in Geiger Counters. so the original posting stands.
I've got my answer regarding the Bickel instrument (Disc or intensity).
However is true that the output of a scintillation tube is pulsed voltage directly related to the initial product of the NaI(T) photons? (photoelectric superimposed in compton).


This is only true in an conceptual idealized isotropic environment. Without going into the subtleties of Shannon's communications theory and simply utilizing practicle real world effects, signal to noise ratios are NOT fixed but independently variable. Either the source may be amplitude varying or the Background noise maybe be amplitude and time varying. In either or both cases the signal to noise ration is affected. If the background noise is of greater amplitude than the source signal, the source is swamped out by the background noise. The result is that the signal to noise ratio changes.

was necessary to mention the (trivial) note, that the source's are constant in time in this case? my intention was to be a simple example.
Now, lets see the more realistic cases.
Any time variance of the sources flux, has effect on the noise to signal ratio but this has nothing to do with any detector.
Let the ratio be variable all the time, with whatever function you like.
The point is that the "ultimate" (barrier) signal to noise ratio is a property of the space it's self, and not instrumentation related.
As you make an instrument more and more sensitive there is a point beyond that any improvement in terms of sensitivity is useless.
This is not because the signal then is buried in noise as in your statement: "..As you increase the sensitivity of a scintillator you increase the level of background radiation proportionately and any signal or signature of interest gets buried in the noise.."
This is simply because, beyond this point you amplify the signal as well the noise - so, no real benefit except signal scaling.


Not false!
sorry, but these "millions of miles of lead shield" was to shield the gamma radiation of a cosmic event in the order of magnitude of the bing bang?


anyway,many thanks for the Bickel info and the nice talking.
this instrument was mysterious, not to say (almost) mythical...
Bill

Geo
02-26-2016, 06:00 AM
After your discussion I feel to be more irrelevant but thanks again:)

zidex666
02-26-2016, 08:39 PM
To Bill512:

See: https://en.wikipedia.org/wiki/Signal-to-noise_ratio

Regards

zidex666
02-26-2016, 10:13 PM
For comparison here are some Photos of a Vintage scintillator (gamma ray detector) unit similar in function and sensitivity the BICKEL unit. The PRI-118 was manufactured and sold during the mid to late 1950's. It utilized sub-miniature vacuum tubes developed during WW2. The appear from time to time on EBAY.

Regards

zidex666
02-26-2016, 10:47 PM
Notice the canister below with the yellow crystal. The rubber O ring seal has leaked allowing moisture into the crystal housing. Crystal units constructed in the 1950 had this problem. They typically lasted only 2 to 3 years. Once yellowed they become useless unless the crystal is removed and replaced in a moisture free environment. Sometimes they can be re-polished. In the 1960' on, manfacturers started to hermetically seal them so that they would last longer.

As mentioned in previous posts by me, NaI(t) is extremely toxic and hygroscopic. It corrodes quickly when exposed to any moisture in the air . Ggreat care and precautions most be made when replacing or re-polishing it once it has yellowed. Once yellowed and no longer transparent, the crystals are useless.

The crystal along with the photomultiplier are the heart of any scintillator unit. The rest are simply electronics that process the output from the photomultiplier tube.

In addition to Nai(t) other materials such as special plastics and other compounds are used for monitoring certain bandwidths of the X-ray/Gamma spectrum are used today.

The myth that only the most sensitive scintillation crystal and/or photomultiplier tubes are only available from Germany is inaccurate and false. Both are now manufactured worldwide by many firms including some in Russia and China.

Regards

Geo
02-27-2016, 06:07 AM
Hello.
Today on ebay we can find scintillation crystals and photomultiplier tubes easy and at low prices. How difficult or easy is to construct a new device like Dr Brickel's ? For example Dr brickel was using a lot of components to make a digital counter and now it is fixed (i mean all components and display at same small case) at very low price. Except the protectiob for good shielding are there other things who makes the construction unposible or unworkable?

Regards

zidex666
02-27-2016, 10:34 PM
Hello.
Today on ebay we can find scintillation crystals and photomultiplier tubes easy and at low prices. How difficult or easy is to construct a new device like Dr Brickel's ? For example Dr brickel was using a lot of components to make a digital counter and now it is fixed (i mean all components and display at same small case) at very low price. Except the protectiob for good shielding are there other things who makes the construction unposible or unworkable?

Regards

Hi Geo,

Yes there are on Ebay. Some really good ones are available from some Russian sellers. I suggest that you get a 6619 photomultiplier tube to mate with a 1 inch diameter NaI(t) crystal as a first project. Another idea is to get a vintage PRI-111 or PRI-111b scintillator or an alternative unit off EBAY and restore it.

Making Bickel's digital counter is unnecessary simply because you can buy panel mounted digital counter modules on EBAY for less than $20 USD. It works kind of like a simple kitchen digital count-down timer. All that counter does is to count "seconds" backwards to shut off an external electromechanical paper chart unit than can be connected to the Bickel unit.

There is simply NO way that the Bickel Counter, no matter what its count display, can accurately identify an individual isotope unless a scalar reading with an external chart recorder is utilized. Even then it's only a guess at best. It merely counts time in Seconds not isotope number. Furthermore, it is extremely improbable that the Bickel unit can identifiy any gold isotope by reading some mysterious number on the counter display. Gold in nature is stable and any gold isotope that is radioactive can only be so if it is "synthetic." That means that it must be strongly irradiated with radioactivity and in essentially laboratory conditions. Additionally any such synthetic radioactive gold isotopes have comparatively short half-lives.

The only things that the Bickel unit along with other NaI(t) based scintillometers detect and are most useful for are
radioactive sources emitting gamma/xray radiation in the field. Primarily decay products from Thorium and Uranium as they eventually decay into lead. Do a search on Uranium and Thorium decay products (example radium. etc.) to learn how Uranium or Thorium eventually transmute into Lead. Also, look up the half-life of the individual elements as they decay into lighter elements in the decay chain.

Finally, there is NO magic or obscure Nazi Science manifest in the Bickel unit. It's a rather conventional Scintillator design with an unnecessary bunch of extra bells and whistles, fancy knobs and unnecessary meters. In order to identify ore fields that were found using the Bickel unit, other extant scintillator at the time could have given equivalent if not superior performance. Without a good knowledge of the geology of the terrain, the out-croppings, and other envronmental factors, which the Bickel may certainly have had, a scintillator will not directly identify oil, gold, dinosaur bones, precious metals, or treasures. The exception maybe if someone buried a nuclear warhead or radioactive uranium mine tailings.

Any NaI(t) based scintillator on the other hand CAN detect long range radioactive (trans-uranium and trans-thorium) ore bodies beneath the surface. Identifying the ore body and its constituents, one can properly identify any other elements such as gold, oil, or other things requires an intimate knowledge of geology, chemical testing, borehole sampling, and other factors in addition to a scintillator.

See below some pictures of common vintage scintilltors we've restored.

FYI: The are some excellent YAHOO interest Groups that are devoted to Geiger Counters and Scintillators.

Regards

Geo
02-28-2016, 06:30 AM
Zidex thank you.
Do you think that one device like these can catch buried gold 100 years at depth 4m and weight 500kgr??

Regards:)

Nicolas
03-01-2016, 04:40 AM
Thanks all friends. you can contact her soon here about more information or inquiries.


http://www.longrangelocators.com/forums/showpost.php?p=73412&postcount=26

abdou2014
03-01-2016, 08:27 AM
Thank You :)

Mike(Mont)
03-01-2016, 02:31 PM
You might try to contact a member here SEDEN. I think he has studied Bickel.

Nicolas
03-01-2016, 11:42 PM
Hi all
Maybe this can help you to build your neutron detectors for long range.

I advice to read it first please to understand good

https://en.wikipedia.org/wiki/Neutron_detection

Second you can try here to find the scheme details.

https://www.google.com/search?q=Neutron+detection&biw=1152&bih=617&source=lnms&tbm=isch&sa=X&ved=0ahUKEwj9kNeyy6DLAhUkGZoKHdypBHIQ_AUIBigB#imgd ii=nZ90ezDnSkmSSM%3A%3BnZ90ezDnSkmSSM%3A%3B1MzBHhK muky8OM%3A&imgrc=nZ90ezDnSkmSSM%3A

Third you can build your neutron detectors here.

http://uzzors2k.4hv.org/index.php?page=rs232geiger_counter

or this

http://www.blackcatsystems.com/GM/geiger_counter.html

or this

http://www.imagesco.com/articles/geiger/build_your_own_geiger_counter_-_gm_tube.html

http://www.berkeleynucleonics.com/wordpress/?m=200910

http://www.berkeleynucleonics.com/products/model_940.html

http://www.berkeleynucleonics.com/images/940-web-training-3.JPG

http://www.berkeleynucleonics.com/wordpress/wp-content/uploads/2009/10/940top-300x245.jpg


http://www.directindustry.com/prod/canberra-industries/product-23661-1346769.html

http://img.directindustry.com/images_di/photo-mg/23661-8601461.jpg

http://www.imagesco.com/articles/geiger/fig2web.jpg

For us we use the thermal detection and we are competent in this domain

https://www.youtube.com/watch?v=oVAlQhLRB1I

https://www.youtube.com/watch?v=uRkqy4K7o9g

https://www.youtube.com/watch?v=Rokwo50EoL0





Good luck to all

Best regards
Nicolas

Nicolas
03-02-2016, 12:10 AM
https://www.orau.org/ptp/collection/surveymeters/surveymeters.htm



https://www.orau.org/ptp/collection/surveymeters/precisionroyal.htm

https://www.orau.org/ptp/collection/surveymeters/precis6.jpg

https://www.orau.org/ptp/collection/surveymeters/precis2.jpg

WM6
03-02-2016, 01:57 AM
Interesting that known metal detector producer White's started his business as Geiger Counter producer,
so practically in Long Range Locater (LRL) field (it even detect "subnormal" phenomenon):

http://national-radiation-instrument-catalog.com/new_page_25.htm

http://i63.tinypic.com/2aaaadx.jpg

Geo
03-02-2016, 06:04 AM
Interesting that known metal detector producer White's started his business as Geiger Counter producer,
so practically in Long Range Locater (LRL) field (it even detect "subnormal" phenomenon):

http://national-radiation-instrument-catalog.com/new_page_25.htm

http://i63.tinypic.com/2aaaadx.jpg


Very interesting.
Also a man "Steve White" is owner of a Bickel device but i don't know if he has any relation with whites electronics. Maybe Carl or someone else knows.

:)

Sneshko
03-02-2016, 06:20 PM
Interesting that known metal detector producer White's started his business as Geiger Counter producer,
so practically in Long Range Locater (LRL) field (it even detect "subnormal" phenomenon):

http://national-radiation-instrument-catalog.com/new_page_25.htm

http://i63.tinypic.com/2aaaadx.jpg


Here's who first discovered the "Phenomenon" ...!
My first metal detector was White`s Coin Master 6000 DI S3.
Obvious, I was a "little late" this White`s LRL model.
Regards!
Sneshko

Bill512
03-02-2016, 07:12 PM
Interesting that known metal detector producer White's started his business as Geiger Counter producer,
so practically in Long Range Locater (LRL) field (it even detect "subnormal" phenomenon):


so, that's why their first metal detector models, had a "Geiger look"....

BENZINAS
03-02-2016, 07:30 PM
The question is can a geiger counter help us find buried gold long time ago ?

WM6
03-02-2016, 09:52 PM
so, that's why their first metal detector models, had a "Geiger look"....

Especially with their "nuclear" Logo:

http://lrlman.com/images/whites3dlogo.jpg


The question is can a geiger counter help us find buried gold long time ago ?

If those gold radiate in Gamma spectrum, why not?

Nicolas
03-02-2016, 10:39 PM
Especially with their "nuclear" Logo:

http://lrlman.com/images/whites3dlogo.jpg




If those gold radiate in Gamma spectrum, why not?



"The gold standard for gamma-ray spectroscopy is the high purity germanium (HPGe) detector," says Fisk Professor of Physics Arnold Burger. "However, it requires cryogenic cooling so it is very bulky. It also needs vacuum-tube technology so it consumes too much energy to run on batteries. SrI2 isn't quite as good HPGe, but it is more than adequate to do the job and it is compact enough and its power requirements low enough so that it can be used in spacecraft and even placed on robotic landers.


=====================
By using a photographic method in connection with semi-circular focusing magnetic spectrometers, the beta-spectra due to internal conversion have been evaluated for gold 198, iridium 194, and tantalum 182. For gold a simple spectrum of 3 lines with differences the same as between the K−L−M levels of mercury indicate a single gamma-ray of energy 0.408 Mev following beta-emission.
In iridium, a complicated spectrum of more than 26 lines is observed which may be identified from the K−L−M differences in platinum as due to 12 gamma-rays. These gamma-energies appear to fit well with a level scheme of six terms.
Tantalum yields a spectrum of about 34 lines from which about 16 gamma-rays may be derived by applying the K−L−M differences for tungsten. These gamma-energies fit satisfactorily on a level scheme of 9 terms.

archive journal

http://journals.aps.org/pr/abstract/10.1103/PhysRev.72.581

pigeon
03-02-2016, 11:00 PM
salam si ridha .

inajem inkalmek 3al 21****49

merci frere

Nicolas
03-03-2016, 05:11 AM
salam si ridha .

inajem inkalmek 3al 21****49

merci frere

Yes but these number is changed
You can send me PM
welcome

Nicolas
03-03-2016, 05:21 AM
http://www.curtcass.com/detectron/index2.html


http://national-radiation-instrument-catalog.com/Detectron%20Ad%201950a.jpg

http://national-radiation-instrument-catalog.com/Cassingham%20Pic%201.jpg


http://www.curtcass.com/detectron/JLC-donkey.jpg


http://www.curtcass.com/detectron/711Schematic.jpg

Schematic for Detectron 711 Metal Detector

http://www.curtcass.com/detectron/711Schematic.jpg

pigeon
03-03-2016, 02:04 PM
Yes but these number is changed
You can send me PM
welcome
salam

your pm is disable.
si tu es ou bled j'aimerai bien te parler si pssible
thank you

Qiaozhi
03-03-2016, 07:46 PM
salam si ridha .

inajem inkalmek 3al 21****49

merci frere

salam

your pm is disable.
si tu es ou bled j'aimerai bien te parler si pssible
thank you
Please read the rules of the forum ->
http://www.longrangelocators.com/forums/showthread.php?t=10526
and make your posts in English.